1) When gathering the ECG signal via AD8233 and ADuCM350, Can we distinguish the position of PQRST wave?
For example, we gathered as below with 3 electrodes mode.
I'm not sure we can distinguish P and T position at below picture, even though we use LPF.
Let me have your opinion.
2) Is there any signal shape difference between left wrist ECG and right wrist ECG ?
Thanks a lot.
Hello.I have an AD8233CB-EBZ evaluation board and I'd like to adapt filter to extract DIAGNOSIS signal (0.05 - 150 Hz).
The AD8233CB-EBZ evaluation board filters are configurated for SPORT (7-25Hz…
Hi Jiwon.Lee,
the three typical bandwidth are:
Regards
Jose
The AD8233CB-EBZ evaluation board filters are configurated for SPORT (7-25Hz).
Can you tell me the values of filter resistor/capacitor R14/C7 - C4/R12 - R13/C8/R16/C9 for adapt AD8233CB-EBZ evaluation board to DIAGNOSIS?
Thank you