Can I do initial calibration of ADATE305 ?
Hi
 Output of  DUT0 /DUT1 is differente when the power supply and code are all the same, Can I do the initial calibration ? And how do it ?
Another, different chips at the same code, same power supply , the output voltage are differente. This also need calibration.
thanks
Thread Notes
  • Hi Aggil,
    One common calibration method you can apply to the ADATE305, is to do a two-point calibration. You have to first get its offset and gain errors. Please see the steps below for manual calibration.
    1. For the offset error, force 0V and measure at Measout pin. The voltage at Measout pin will be the offset error.
    2. For the gain error, force 2 values (e.g. 2V and 5V) and measure these voltages at Measout pin. Then, subtract the values of the two and divide it by the difference of the two forced voltages. The result will be your gain error.
    3. On your software, you have to subtract the offset error to your programmed value and then, divide it to the gain error. The value you get will be now your new programmed value or your calibrated value. Below is the equation.
                 Calibrated value = (Programmed Value – Offset Error)/(Gain Error)
    Best Regards,
    DJ
  • Hi
      Thank you.
      Another, As below table.
    • The actual output voltage of VL differs greatly from the theoretical value, and the error is > 700MV
    • The actual test of VT/VH/VL upper and lower voltage has a large error with the theoretical value of the upper and lower voltage
    • HVOUT has a maximum voltage output of only 15.13V and a minimum of -0.68V. It is different with ds.
    IC NO. Drive state Reg add set voltage code The theory of voltage(V) practical DUT0(V) practical DUT1(V) HVOUT(V) 备注
    PEC32_4#
    ADATE305_1#
    VL 0X02 0X2ccc 4.5 3.9 3.85
    0X3fff 7.5 6.36 6.39
    0X0 -2.5 -2.35 -2.37
    VH 0X01 0X2ccc 4.5 4.52 4.5
    0X3fff 7.5 7.06 7.1
    0X0000 -2.5 -1.24 -1.26
    VT 0X03 2ccc 4.5 4.53 4.51
    3fff 7.5 7.08 7.09
    0X0000 -2.5 -2.36 -2.39
    PMU 0X0b 0Xb333 4.5 4.48 4.49
    ffff 7.5 7.47 7.49
    0X0000 -2.5 -2.513 -2.5
    VHH 0X03 2ccc 11     11.1 VHH = (VT + 1 V) × 2
    3fff 17     15.13
    0X0000 -3     -0.68
    VDD = +10.17 V, VCC = +3.3 V, VSS = −5.68 V, VPLUS = +16.47 V, VCOMP_VTT = +3.3 V, VREF = +4.98 V, VREF_GND = 0.0 V
  • Hi Aggil,
    Can we take this offline? Please send an email to me at danilojames.marciano@analog.com.
    Best Regards,
    DJ
  • Hi
      thanks ,we wil summarize the question and mail to you.