AD8672 working temperature
Product Number: AD8672
We’re using an AD8672 as both a gain/shift buffer and an ADC input driver, powered by ±12 V rails.
In a “standby” condition, the AD8672 package temperature measures around 40+ °C—roughly 10 °C higher than other op amps (e.g., OPx177) on the same board under similar conditions.
To investigate, we tried:
  1. Letting the input (Vin) float or tying Vin to ground.
  2. Removing R10 and R16 to float the output.
In both cases, the device temperature remains around 40+ °C. We’ve attached our schematic and a temperature diagram for reference.
Is this temperature rise typical for an AD8672 operating at ±12 V? Are there any design considerations or power dissipation factors we should check to confirm this is within normal operating limits?
Thanks for your assistance!
  • Hello Cain,
    Does this happen on other boards too? Have you tried another AD8672 part to check if it is not part specific?
    Thanks and Best Regards,
    Kenneth
  • Hi Kenneth,
    Yes, this happens on all our prototype boards, not just a single unit or a single AD8672 part.
    Our AD8672’s input (Vin) is driven by a MUX, so initially, we suspected that frequent switching of the MUX might be adding extra power consumption. However, even when we keep the MUX on a fixed channel or disconnect the output, the AD8672 still settles at around 40°C.
    We’re mainly curious if this behavior is considered “normal” for the AD8672 operating on ±12V rails. Any insight or guidance would be greatly appreciated.
    Kind regards,
    Cain
  • Hello  ,
    Sorry for the late reply. I checked this query again and upon looking at the DS, it seems that what you are seeing is correct for AD8672.
    On page 11 on the DS at Power Dissipation section, it says junction temperature is expected to be 42C if ambient temp is 30C on AD8671 single MSOP package.
    Since you are using AD8672, power dissipation is twice as that of single package, so you will get around 23.8C rise above ambient temperature.
    So let's say your ambient temp is 23C, you'll expect 46.8C junction temp when both opamps are powered up (rise in temp is excluding load currents on your output if your op amp is connected to a load).
    Kenneth