Hi,
for RTD measurement I use the current source from output pin 6 "IOUT_HV" and excitate with it RTD and 500 Ohm precise reference resistor in series. With the PGA ADA4254 I measure
the voltage drop through RTD on CH1 (IN1+/IN1-) and the voltage drop through the reference resistor on CH2 of the same PGA (IN2+/IN2-) for ratiometric measurement. I do not measure the
voltage drop at the same time, because I need to switch from CH1 to CH2 measurement. In this two time stamps the value of the excitation current is not the same, because of noise or other influences, so I get
an error in resistor measurement, because of the instability of the excitation current during time. In an older application I sourced the RTD measurement from ADC AD7124 and had more stability
in excitation current as from the PGA excitation current source. I tested with 4kOhm RTD and had an variation of dR=0.53 Ohm with current source from ADC and dR=1.55 Ohm with current source from PGA.
This precision is not enough for my application.
Do you know about the instability of the excitation current source of the PGA ADA4254 and why is the current source from ADC ADA4254 more stable?
Have you recommandations to reduce the instability of the current source or other current source you know that have comparible stability as the ADC?
Could you tell which kind of current source circuit you are using in PGA and ADC which maybe explains the difference of the stability. I could not found values in the datasheet of both chips.
Thank you in advance,
Thomas