AD8253 excessive bias current
Using AD8253 in classic scope-type DAQ front end. Both inputs have 1 Mohm to ground. In an AC-coupled mode, the amp Ib develops offset voltage across the input resistor. Data sheet specs indicate 5 nA typical, 50 nA max over temp, further supported by Figures 8, 9, 18 and 19. I am seeing consistent 30-50 nA = 30-50 mV on tens of samples with all my other circuitry isolated. Any hints?
  • Paul,
    Thank you for the feedback. I see the same amount of offset when I
    remove all ancillary components and have only a 1M resistor from AD8253
    pins 1 and 10 to AGND, so the Ib and Ios both seem to be running toward
    the maximum spec on almost all the chips I've tried.
    The ideal DC reading is as close to zero millivolts as possible, so the
    target has to be pragmatic. In practice users have found 5-10 mV
    acceptable but 30-60 mV is a  noticeable and troublesome offset for the
    typical AC-Coupled IEPE accelerometer application. I don't have room or
    power for a discrete FET-input in-amp topology in this design. If that
    is looking like the norm for the AD8253 part at +/-7V I'll have to
    figure out some other mitigation strategy. I have to say it makes the
    "typical" numbers published in the datasheet not only a waste of ink as
    Harry pointed out but actively misleading. I appreciate that you folks
    have always been helpful and frank even when the answer is not what I
    would have wanted.
    Gary
    on Aug 2, 2018 3:45 PM
    Thank you,
    EZ Admin