ADALM1000 impedance analyzer calibration

Thread Summary

The user is experiencing incorrect capacitance measurements with the ADALM1000 impedance analyzer, particularly in the 1nF to 1uF range. The support engineer suggests that the factory calibration should be sufficient and recommends reviewing the 'Impedance Analyzer / LCR Meter' section of the user guide for examples. For small capacitance values, the parasitic capacitance of the M1k (~400pF) can affect measurements, and an open circuit test followed by clicking the Zero button under Capacitance Offset can help null out this effect.
AI Generated Content
Category: Software
Product Number: ADALM1000
Software Version: 1.3.14.29

Hi,

I am trying to use ADALM1000 impedance analyzer.

It seems working correctly if I measured resistor. But if I measured capacitor, its value seems not correct.

I tried to perform calibration with:

https://wiki.analog.com/university/tools/m1k/alice/desktop-sa-ba-ia-users-guide

its expression is just one example and not include actual way of calibration in general.

Would you suggest me how to calibrate for measuring capacitor?

I wanted to measure 1nF to 1uF region.

Regards,

Satoru

Parents
  • You shouldn't need to do any (DC) calibration, the factory calibration should be good enough.

    Have you tried the example in the "Impedance Analyzer / LCR Meter" section of this page on the Wiki?

    wiki.analog.com/.../desktop-sa-ba-ia-users-guide

    To help you use the AC impedance analyzer virtual instrument I'm going to need some additional details on your setup.

    First, what range of capacitance are you trying to measure? Very small values will be swamped by the input capacitance of the M1k (~400pF). Second what value of external test resistance are you using? Again you need to pick a resistor value such that the RC time constant of the cap under test and the resistor is reasonable when compared to the test frequency. Which I guess is the third question, at what frequency are you testing the cap? In general the measurement will only be as accurate as the known (value entered) Rext resistor you are testing against.

    For small values of capacitance, around the M1k input capacitance, you can "Null Out" the parasitic by doing an open circuit test as it were (no cap connected) and clicking on the Zero button under the Capacitance Offset heading. 

Reply
  • You shouldn't need to do any (DC) calibration, the factory calibration should be good enough.

    Have you tried the example in the "Impedance Analyzer / LCR Meter" section of this page on the Wiki?

    wiki.analog.com/.../desktop-sa-ba-ia-users-guide

    To help you use the AC impedance analyzer virtual instrument I'm going to need some additional details on your setup.

    First, what range of capacitance are you trying to measure? Very small values will be swamped by the input capacitance of the M1k (~400pF). Second what value of external test resistance are you using? Again you need to pick a resistor value such that the RC time constant of the cap under test and the resistor is reasonable when compared to the test frequency. Which I guess is the third question, at what frequency are you testing the cap? In general the measurement will only be as accurate as the known (value entered) Rext resistor you are testing against.

    For small values of capacitance, around the M1k input capacitance, you can "Null Out" the parasitic by doing an open circuit test as it were (no cap connected) and clicking on the Zero button under the Capacitance Offset heading. 